Questioning the classification of “high blood flow” versus “low blood flow” ECCO₂R in ultra-low tidal volume ventilation studies: a call for functional classification

Matters Arising to this article was published on 18 April 2025

The Original Article was published on 27 December 2024

Dear Editor:

We read with great interest the recent study by Monet et al. [1] investigating the feasibility and safety of ultra-low tidal volume ventilation (≤ 3 mL/kg) combined with ECCO₂R in acute respiratory failure. While the study provides valuable insights into lung-protective strategies, we wish to highlight a critical point in the authors' classification of ECCO₂R devices into “high blood flow” (HBF) and “low blood flow” (LBF) groups, which may undermine the validity of their conclusions.

Arbitrary thresholds compromise comparability

The study defines HBF as “blood flow ≥ 1000 mL/min” and LBF as “blood flow < 500 mL/min” without citing standardized criteria. This dichotomy ignores two key issues:

Threshold variability: existing literature uses conflicting cutoffs (e.g., HBF as > 800 mL/min in SUPERNOVA study [2]).

Functional disconnection: blood flow alone poorly predicts CO₂ clearance. For example, the Prismalung + ®(classified as LBF in the study) achieves 90 mL/min above CO₂ removal at 400–450 mL/min blood flow [3], surpassing some “HBF” devices at 500 mL/min with smaller membrane surfaces.

By prioritizing blood flow over CO₂ extraction rate (mL/min) and membrane efficiency (CO₂ clearance per L blood flow), the authors risk misclassifying device performance. A device with 500 mL/min above flow but low membrane efficiency may be functionally inferior to a 400 mL/min device with optimized design, yet both would be grouped differently in this analysis.

Confounding by device heterogeneity

The HBF/LBF grouping aggregates fundamentally distinct technologies: HBF group includes pumpless arteriovenous devices (e.g., iLA Activve®) while LBF group combines with roller pump system (e.g., PrismaLung +®) and centrifugal pump system (e.g., Hemolung Respiratory Assist System®), ignoring their divergent hemodynamic impacts [4].

This heterogeneity introduces unmeasured confounding. For instance, the reported “no significant safety difference” between groups could mask device-specific risks (e.g., hemolysis in centrifugal pumps vs. thrombosis in pumpless systems).

Overlooked interactions with ventilation strategy

The study’s primary endpoint—feasibility of ultra-low tidal volume ventilation—depends on precise CO₂ control, which is determined by ECCO₂R efficiency (CO₂ clearance/mL blood flow), not absolute flow rates. A functional classification based on CO₂ extraction capacity would possibly better predict the ability to maintain pH and PaCO₂ targets.

Therefore, to advance future ECCO₂R research, we propose:

Standardized functional metrics: report CO₂ extraction rate (mL/min) normalized to blood flow (mL/min) and membrane surface area (m2).

Device-specific subgroup analyses: compare outcomes by technology type (e.g., centrifugal vs. roller pump systems) rather than arbitrary flow categories.

Dynamic performance assessment: incorporate real-time CO₂ clearance data during dose titration, as static flow thresholds cannot capture device responsiveness to metabolic demands.

While Monet et al. [1] contribute importantly to the field, re-evaluating ECCO₂R classification strategies is essential to avoid misleading conclusions and guide evidence-based device selection.

Availability of data and materials

No datasets were generated or analysed during the current study.

Abbreviations

ECCO₂R:

Extracorporeal carbon dioxide removal

HBF:

High blood flow

LBF:

Low blood flow

References

  1. Monet C, Renault T, Aarab Y, et al. Feasibility and safety of ultra-low volume ventilation (≤ 3 ml/kg) combined with extra corporeal carbon dioxide removal (ECCO2R) in acute respiratory failure patients. Crit Care. 2024;28:433. https://doiorg.publicaciones.saludcastillayleon.es/10.1186/s13054-024-05168-8.

    Article  PubMed  PubMed Central  Google Scholar 

  2. Combes A, Tonetti T, Fanelli V, et al. Efficacy and safety of lower versus higher CO2 extraction devices to allow ultraprotective ventilation: secondary analysis of the SUPERNOVA study. Thorax. 2019;74(12):1179–81. https://doiorg.publicaciones.saludcastillayleon.es/10.1136/thoraxjnl-2019-213591.

    Article  PubMed  Google Scholar 

  3. Hospach I, Goldstein J, Harenski K, et al. In vitro characterization of PrismaLung+: a novel ECCO2R device. Intensive Care Med Exp. 2020;8(1):14. https://doiorg.publicaciones.saludcastillayleon.es/10.1186/s40635-020-00301-7.

    Article  PubMed  PubMed Central  Google Scholar 

  4. Gross-Hardt S, Hesselmann F, Arens J, et al. Low-flow assessment of current ECMO/ECCO2R rotary blood pumps and the potential effect on hemocompatibility. Crit Care. 2019;23:348. https://doiorg.publicaciones.saludcastillayleon.es/10.1186/s13054-019-2622-3.

    Article  PubMed  PubMed Central  Google Scholar 

Download references

Acknowledgements

None.

Funding

None.

Author information

Authors and Affiliations

Authors

Contributions

MW designed and wrote the manuscript. QY & MZ reviewed it.

Corresponding author

Correspondence to Minmin Wang.

Ethics declarations

Ethics approval and consent to participate

Not applicable.

Consent for publications

Not applicable.

Competing interests

MW&QY declare to have competing interests. MZ declare to have no competing interests. This article is not supported by any company or funding.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution-NonCommercial-NoDerivatives 4.0 International License, which permits any non-commercial use, sharing, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if you modified the licensed material. You do not have permission under this licence to share adapted material derived from this article or parts of it. The images or other third party material in this article are included in the article’s Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article’s Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by-nc-nd/4.0/.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Wang, M., Yao, Q. & Zhu, M. Questioning the classification of “high blood flow” versus “low blood flow” ECCO₂R in ultra-low tidal volume ventilation studies: a call for functional classification. Crit Care 29, 121 (2025). https://doiorg.publicaciones.saludcastillayleon.es/10.1186/s13054-025-05352-4

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doiorg.publicaciones.saludcastillayleon.es/10.1186/s13054-025-05352-4

Keywords